13 GATE previous year questions on Ultrasound Imaging (Medical Imaging Systems, Biomedical Engineering) with answers and explanations, from every paper.
- GATE 2020 BM Q17 – A 5 MHz ultrasound pulse images a tumor 2 cm deep in soft tissue. Its round-trip echo time is t. If frequency changes to 2.5 MHz, what is the echo…
- GATE 2020 BM Q52 – A 3 MHz transducer sends a three-cycle pulse normally toward a fat–liver boundary. Use tissue speed 1500 m/s, fat speed 1450 m/s,liver speed 1570 m/s,…
- GATE 2021 BM Q51 – A longitudinal ultrasound wave meets a muscle–kidney boundary at 60°. Acoustic impedances are 1.70×10⁵ and 1.62×10⁵ g cm⁻² s⁻¹; sound speeds are 1590…
- GATE 2021 BM Q65 – A PZT crystal has thickness 1 mm and sound speed 4000 m/s. Its longitudinal ultrasound beam meets blood flow at 30°. Blood speed is 10 cm/s and tissue…
- GATE 2022 BM Q34 – A PZT5A ultrasound transducer must resonate at 5 MHz. Sound speed in the crystal is 4350 m/s. Find the required thickness in mm to three decimals.
- GATE 2023 BM Q35 – A 5 MHz ultrasound transducer measures blood velocity using an echo frequency shift of 200 Hz. The beam makes 45° with blood flow, and sound speed is…
- GATE 2023 BM Q45 – An ultrasonic wave of incident pressure amplitude P₀ strikes a boundary normally from acoustic impedance Z₁ toward Z₂. What is the total pressure…
- GATE 2024 BM Q34 – A 5 MHz Doppler probe measures maximum shift 3 kHz at 45° to blood flow. Sound speed in blood is 1500 m/s. Find maximum flow velocity to two decimals…
- GATE 2024 BM Q55 – A circular ultrasound transducer has radius a=8 mm,frequency 1 MHz,and axial pressure P(r,0)(ka2/(4r)). Sound speed is 1600 m/s. Find positive…
- GATE 2025 BM Q43 – Which statements about ultrasound are true?
- GATE 2026 BM Q22 – Which is the correct decreasing order of ultrasound propagation speed?
- GATE 2026 BM Q45 – An ultrasound wave is normally incident from muscle onto bone. Their acoustic impedances are 1.710-4 and 6.110-4 kg m⁻² s⁻¹ respectively, as printed…
- GATE 2026 BM Q46 – An ultrasound beam travels 5 cm through material attenuating it by 2 dB/cm. Find exit intensity divided by entry intensity.