GATE 2022 BM – Question 29
Which statements about pulse oximetry are correct?
Practise this question in The GATE Grind →
Show answer and explanation
Correct answer: (A) Arterial SpO₂ can be estimated noninvasively from transmitted light and the systolic pulsatile blood component; (B) An isosbestic wavelength gives equal absorbance for Hb and HbO₂; (C) SpO₂ can be determined using the absorbance ratio at 660 and 905 nm
Explanation
A identifies isolation of the arterial pulsatile component. B is the definition of an isosbestic wavelength. The source accepts red 660 nm combined with near-infrared 905 nm as its two-wavelength pair, C, with appropriate empirical calibration. Both wavelengths in D are on the near-infrared side and do not provide the intended red/infrared discrimination. A,B,C are the official selections; a practical oximeter uses calibrated pulsatile ratios, not an uncalibrated raw optical ratio.