An EEG PSD has band-average values (μV²/Hz): 0.4–4 Hz:3; 4–8:4; 8–13:8; 13–30:5; 30–100:3. The noise floor is 2 μV²/Hz. Find the alpha-band SNR in dB using the source’s band-PSD/noise-floor convention.
The alpha band is eight to thirteen hertz and its quoted PSD is eight. Dividing by the noise floor two gives power ratio four; common bandwidth cancels. Therefore SNR=10 log₁₀4≈6.02 dB, C. This is the published convention using the quoted band PSD directly, rather than first subtracting the noise floor.