GATE 2024 IN – Question 21

Sensors and Industrial Instrumentation · Temperature and analytical sensors · 1 mark · Numerical answer

A linear transducer is calibrated over the ranges shown: input temperature 50–150 °C and output current 4–20 mA. Its gain is ______ mA/°C (two decimal places).

Diagram for GATE 2024 IN question 21

Practise this question in The GATE Grind →

Show answer and explanation

Correct answer: 0.15 to 0.17

Explanation

The input span is $150-50=100$ °C; the output span is $20-4=16$ mA. The linear slope is $16/100=\boxed{0.16}$ mA/°C. The 4 mA live-zero offset does not enter the gain.